Originally posted by macona
View Post
Remember, on ANY scope, analog or digital, if you get close to the frequency limit with the fundamental, you filter off the harmonics. So, if you look at a 20 MHz signal with a 100 MHz 'scope, you see only the first 4 harmonics, the rest are filtered off.
When you think about it that way, you quickly see that a 100 MHz 'scope is not an extravagance at all. And the sample rate is easily seen to be important. A 1GHz sample rate translates to a cutoff which cannot be higher than 500 MHz, but at that frequency, you get only one sample..... with 100 MHz, you now get 5 samples per half cycle, which allows at least some semblance of detail. In reality, the 1 GHz sample rate provides detail similar to an analog 'scope up to around 30 MHz, maybe less.
Of course, the "sample rate" of the analog scope is related to the cutoff frequency. And it will lie to you by showing you a nice continuous waveform, with a slight bump on it. That "bump" is actually your evidence of a big transient at a frequency significantly above the cutoff frequency.
The good news is that you saw it. And that with the analog 'scope, the cutoff is typically more gradual, so things can still be "seen", although the amplitude cannot be trusted in any way, above cutoff. With the digital 'scope, you probably would not have seen that transient at all.